Duty factors for imaging systems must always be ______.

Sharpen your skills for the Davies Publishing SPI Test with targeted flashcards and multiple-choice questions, complete with hints and clarifications. Prepare thoroughly for success!

Multiple Choice

Duty factors for imaging systems must always be ______.

Explanation:
Duty factor is the percentage of time the X-ray beam is on during one cycle of operation. Imaging systems are designed to deliver radiation in pulses to limit patient dose and heat load on the tube, so the beam is on only part of each cycle. That means the on-time divided by the total cycle time is less than 1, i.e., less than 100%. For example, if the beam is on for 10 ms in a 60 ms cycle, the duty factor is 10/60 = 16.7%. Therefore, the best answer is less than 100%.

Duty factor is the percentage of time the X-ray beam is on during one cycle of operation. Imaging systems are designed to deliver radiation in pulses to limit patient dose and heat load on the tube, so the beam is on only part of each cycle. That means the on-time divided by the total cycle time is less than 1, i.e., less than 100%. For example, if the beam is on for 10 ms in a 60 ms cycle, the duty factor is 10/60 = 16.7%. Therefore, the best answer is less than 100%.

Subscribe

Get the latest from Examzify

You can unsubscribe at any time. Read our privacy policy